Questions tagged [chromatography]

collective term for a set of laboratory techniques for the separation of mixtures

Filter by
Sorted by
Tagged with
2
votes
1answer
52 views

The best chromatographic technique to analyse a solution of benzene, toluene, and ethylbenzene in n-hexane

What is the best chromatographic technique to analyse a solution containing benzene, toluene, and ethylbenzene in n-hexane? Is gas chromatography (GC) the best way to analyse this solution? Or is ...
3
votes
1answer
129 views

Types of chromatogram shapes [closed]

I'm trying to generate a testing set of different chromatograms to check peak detection (whether peak boundaries are detected correctly) and smoothing algorithms. Currently I'm figuring out which ...
1
vote
0answers
26 views

Column Dispersive Mass Balance analytical / numerical solution

Is there any analytical solution for the most common expression of column mass balance: $\dfrac{\partial C}{\partial t} + \left(\dfrac{1+\varepsilon}{\varepsilon}\right)\dfrac{\partial q} {\partial t} ...
4
votes
1answer
165 views

Stronger ESI signal for deuterated substances

I observe a constant phenomenon in my ESI measurements, which indicates that deuterated substances give a stronger signal than the same substance in non-deuterated form. Unfortunately I can't make any ...
0
votes
1answer
68 views

Is there any database, or reference, for retardation factors (Rf)? [closed]

I'll like to compare the Rf of myrcene, obtained in Thin Layer Chromathography, for different species. Toluene:Ethyl Acetate (97:3) would be fine, but is not a necessity.
1
vote
1answer
52 views

Theoretical plates in Plate theory of chromatography

In "Chromatography theory" by Scott and Cazes, The plate volume is given as ($v_m+Kv_s$) where $v_m$ and $v_s$ are the volume of stationary and mobile phase in the column and $K$ is the ...
2
votes
0answers
35 views

How do I analyze 17 different gases from a sample?

I am designing a test setup in which I heat a battery until it releases its gases, and I need to analyze the composition of the fumes. Below is the list of gases that are believed to be present in the ...
-1
votes
1answer
28 views

Elution and void volumes in a gel chromatography experiment

I was using gel-filtration for separation Hb from xylene cyanol FF with Blue Dextran to determine the void volume. Sephadex G-25 is stationary phase, and mobile phase was buffer $(\pu{0.020 mol dm-3}$ ...
1
vote
0answers
23 views

Disadvantages of SIM vs TIC in GC-MS [closed]

In GC-MS what are the relative advantages/disadvantages of using Selected-Ion Monitoring (SIM) vs using Total Ion Chromatograms (TIC)? I know that SIM offers a better S/N ratio, however if the S/N is ...
1
vote
0answers
58 views

Purification of primary amines using Schiff base immobilization

I want to purify all primary amine containing compounds in a complex mixture (cell extracted metabolites). I am not aware of any published methods to do so, so I am thinking to make my own using ...
3
votes
1answer
81 views

How do I import CSVs to OpenChrom?

My end objective is to compare GC outputs of samples in two groups using PCA. I have bene given the data in Excel format but have saved to CSV. Whilst I can locate CSV files through OpenChrom, they ...
0
votes
1answer
36 views

What order would these compounds elute in a 3:1 hexanes:ethyl acetate TLC run? [closed]

I would figure the order would be (in order of smallest Rf to largest) W, Z, Y, X. Is the carboxylic acid more polar than the acetic anhydride?
2
votes
1answer
49 views

Why can zwitterions be difficult detect by HPLC-MS?

I understand that to separate molecules by HPLC (or SFC) you commonly adjust the pH of the mobile phase to get preferable charges (i.e. negative charges) on the molecules - so that they interact with ...
3
votes
1answer
81 views

Is there a list of must-have columns for LC-MS/MS used in biochemistry/clinical lab setting?

Considering a wide variety of chromatographic columns used for (U)HPLC based on phase type (normal or reverse); column dimensions (geometry, length, inner diameter); particle material $(\ce{SiO2},$ $\...
-1
votes
2answers
173 views

What is the principle of RP-HPLC? [duplicate]

Is the following statement the correct way to explain principle of RP-HPLC? It works on the principle of hydrophobic interactions hence the more nonpolar the material is, the longer it will be ...
1
vote
2answers
108 views

Can separated and purified substances be obtained from a chromatogram?

I am studying chemistry, and I am learning about chromatography. But I am confused about how chromatography is used to ‘separate and purify’ substances. I have learnt about chromatography paper, ...
2
votes
1answer
95 views

Conditions for choosing quantifying and qualifying ion in LC-MS/MS

In my PhD-work, we are trying to develop LC-MS/MS methods for targeted analysis. Usually, I determine my quantifying ion/fragment based on a few different criteria: If there is a 13C internal ...
5
votes
1answer
81 views

How to recover C18 column after prolonged exposure to phosphate buffer?

A colleague of mine who occasionally does HPLC in a biochemistry lab discovered that a C18 column was not properly flushed after the previous experiment. It appears that the column was stored for ...
2
votes
1answer
60 views

Can we use HPLC to purify an organic reaction product?

I know HPLC is a form of column chromatography that uses high pressure to pump a sample mixture/eluent through a column. However, I'm not sure if this technique is being used to purify and isolate ...
2
votes
1answer
109 views

Software platforms with integration of multi vendor mass spectrometer device data?

I know about ACD/Spectrus Processor which is a software platform to integrate analytical chemistry data from multiple mass spectrometer device vendors (e.g. Agilent Technologies, Bruker, Waters) to ...
2
votes
1answer
50 views

Why can DAP be used to elute a protein in affinity chromatography?

A little background: as part of my bioinformatics degree I have to take Protein Chemistry course, but I miss chemistry basics (have CS background), so that's why I'm asking the following question. We ...
2
votes
1answer
90 views

Why n-alkanes can be found in organic solvent extract of rubber?

Sorry for the confusing title, I didn't know how to describe this better. So, if we douse rubber, say, a bicycle tire or a certain type of rubber glove, with organic solvent (n-heptane), we can find n-...
3
votes
2answers
512 views

What dictates cathode vs anode nomenclature use?

what was the justification for the use the cathode and anode terminology in IEF? There is no redox taking place in IEF, just proteins interacting with an electric field
0
votes
1answer
61 views

When to use chromatographic separation vs vacuum-assisted rectification distillation?

I'm trying to understand a bit more about industrial scale separation processes. One particular element I'm interested in is the question as to when chromatography (HPLC or Simulated Moving Bed) may ...
0
votes
1answer
51 views

Why mustn't the sample be dipped in the mobile phase in paper chromatography?

As the title says it, I am curious about why the sample (ie, an extract of plant's chlorphyl transferred onto paper) must not be dipped into the mobile phase (ie, the solvent which goes up the paper)? ...
6
votes
3answers
550 views

In High Performance Liquid Chromatography, why are ratios of solvents used?

How do they allow for better separation? For example, why isn't 100% hexane used?
1
vote
1answer
55 views

Organic solvent for cleaning autosampler syringe

I work in a laboratory where gas chromatography is used extensively. We handle phthalates. What are the possible wash solvent for autosampler syringe? I've been using acetone.
3
votes
1answer
128 views

Choosing LC-MS modifier - why only acids?

During method selection our chromatographists have options like NH4, TFA, Formic Acid. What's the point of choosing only between acid? Why not use the strongest that guarantees ionization? The sample ...
1
vote
2answers
60 views

Can I normalize my data in order to compare it?

I have three chromatograms of the same eluted protein but at different concentrations, since it got diluted during purification process) and hence peak maximums at very different absorbances. The ...
2
votes
1answer
205 views

How to pool fractions after chromatography?

I used chromatography to purify my protein and now I have fractions with different concentrations ranging from 1-16 mg/ml. Should I pool all these fractions or just the ones with high concentration (...
1
vote
0answers
18 views

I am having difficulties with analysing steroids in GC/MS [closed]

Is there anyone here that has already analysed estrone and 17beta-estradiol on GC/MS (HP5-MS column and MSTFA as derivatizing agent)? Is it necessary to make a mixture of other reagents with MSTFA or ...
3
votes
1answer
40 views

How appropriate/adaptable is HPLC for in situ sample analysis on Titan?

I have a limited background in chemistry (nothing beyond general Chem 2), so I am asking for help with a group project. We have two chemists on our team, but communication is limited at the time and I ...
2
votes
0answers
19 views

Do I have to change my sample buffer before IEX chromatography?

I will use IEX chromatography to purify a protein and will then use a equilibration buffer Tris-HCl, pH 8 to set the start conditions for my column and optimize the binding of my protein. Then inject ...
1
vote
0answers
21 views

Change buffer in dialysis?

I have a protein that I purified in PBS buffer, pH 7. I will do dialysis to remove salt and will then further purify the protein with ion exchange chromatography. I will need to use another buffer (...
1
vote
0answers
33 views

Protein purification pressure limits confusion, about the X-MPa limit for FPLC?

I have always been a bit confused about column pressure limits. I have to elaborate a bit on the problem to get to my question: In protein purification we use the äkta purification machines (e.g. ...
1
vote
1answer
95 views

Column length for baseline separation

I have a GC-chromatogram with 4 retention times and peak widths and a column length. How can I calculate the column length that I need in order to get baseline separation for all components in the ...
1
vote
0answers
30 views

Acid and base separation on an anion exchange column [closed]

I'm new with this concept and so I would need some advice. I want to separate an acid and a base in a water solution. The acid has $\mathrm{p}K_\mathrm{a}(\ce{HA}) = 5$ and the corresponding acid of ...
3
votes
3answers
98 views

Applying vacuum at the end of a flash chromatography column

I know that flash chromatography uses a positive or pushing pressure to achieve the goal which is separation. Would the same goal be achieved if you had a vacuum on the other end instead of a ...
1
vote
1answer
239 views

In HPLC, how does column diameter affect the retention time?

I am using a smaller diameter column than my previous one keeping all other factors such as stationary phase, solvent, solvent gradient constant. (Obviously, flow rate will be less in smaller column). ...
1
vote
1answer
73 views

Physical meaning of tm

The definition of dead time ($t_\mathrm m$) is mentioned as the time taken for elution of an unretained species. It is also mentioned that the retention time ($t_\mathrm r$) is sum of the time analyte ...
0
votes
2answers
3k views

Retention time and dead time

Why does ts start just after tm ? Does the analyte starts spending time in the stationary phase only after the mobile phase has passed and reached the detector ? Wouldn't there be an overlap between ...
0
votes
1answer
1k views

Concept of Adjusted Retention Time in Chromatography

Why is the relation between the adjusted retention time, retention time and dead time as follows : tR' = tR - tM ? To find the time that the sample spends in stationary phase (adjusted retention ...
4
votes
2answers
230 views

Gas chromatography flame ionization detector (FID) - why hydrogen gas?

Why is $\ce{H2}$ used in an FID, apart from the fact its combustion does not contaminate the flame? In other words, is the temperature of the flame important or critical? Is the stability of the ...
2
votes
1answer
171 views

How to determine polarity of components in TLC

Estimate how a TLC-analysis would look like for the following reaction when you take a TLC at: a) the start of the reaction, b) after 65% conversion and c) after full conversion of A. (You may assume ...
2
votes
1answer
107 views

Use a solid or liquid stationary phase in gas chromatography

In my textbook, they mention that the stationary phase used in gas chromatography is a solid, such as silica or an adsorbed high-boiling-point liquid. My question is that what is the advantage of ...
2
votes
0answers
86 views

ApexTrack: peak tail looks like a shoulder

I'm trying to implement ApexTrack algorithm in my software by following the description written by Waters. But many chromatograms that I see have noticeable tailing of peaks: . No matter what ...
7
votes
1answer
638 views

What is the purpose of trimming the bottom corners of TLC plates?

During my chemistry lab course, I was told by my instructor to make 2 small cuts on the bottom corners of the TLC (Thin-Layer Chromatography) plate. Does this make the solvent rise uniformly on the ...
1
vote
2answers
88 views

In Ion Chromatography instrument, Does the specific conductivity (peak area) of an analyte depends upon background conductivity?

In Ion Chromatography instrument, Does the specific conductivity (peak area) of an analyte depends upon background conductivity ? Assume Specific Conductivity (peak area) of fluoride ion is 4 unit ...
-4
votes
2answers
276 views

Pre-Hydrogen Feature On the Output Of a Gas Chromatography Thermal Conductivity Detector

A candidate for "dark matter", predicted and now apparently detected to have both a higher thermal conductivity and higher diffusion rate than H2 is purported in this PowerPoint Presentation file. ...
0
votes
0answers
44 views

Monitoring amino acid esterification

I am trying to monitor the esterification of a non UV active amino acid.I used thionyl chloride and Methanol as reagents. As amino group is unprotected, Ninhydrin should give positive result for both ...