2
$\begingroup$

A steel sample ($\pu{0.3500 g}$) was dissolved in nitric acid and the resulting solution made up to $\pu{500 ml}$. The absorbance of this solution measured by atomic absorption was found to be $0.410$. A standard $\pu{6 ppm}$ solution of $\ce{Ni}$ gave an absorbance of $0.522$ under the same conditions. Assuming linearity between absorbance and concentration, calculate the $\% \ce{Ni}$ in the steel sample. (8 marks)

Seems so simple but is confusing me.

A steel sample ($\pu{0.3500 g}$) was dissolved in nitric acid and the resulting solution made up to $\pu{500 ml}$.The absorbance of this solution measured by atomic absorption was found to be $0.410$.

A standard 6 ppm solution of Ni gave an absorbance of $0.522$ under the same conditions.

In the final diluted sample, $$\frac{\pu{6 ppm}\ \ce{Ni}}{0.410\times0.522}=\pu{28.03476311 ppm}\ \ce{Ni} \tag{1 }$$

In the first solution $$\frac{\pu{28.03476311 ppm}\ \ce{Ni}}{\pu{500 ml}}=\pu{0.05606952622 mg}\ \ce{Ni} \tag{2}$$

Supposing the density of the first solution to be near that of water:

$$\pu{1000 g}\times\pu{0.05606952622\times10^{-6} g}=\pu{5.60695\times10^{-5} g} \tag{3}$$

In the steel sample:

$$\frac{\pu{5.60695\times10^{-5} g}}{\pu{0.35 g}}=0.000160198=0.016\ \%\ \ce{Ni}\tag{4}$$

$2$nd attempt:

\begin{align} \frac{\pu{6 ppm}}{0.522}&=\frac x{0.410}\\ \implies \frac{\pu{6 ppm}\times0.410}{0.522}& =x\\ x&=\pu{4.7126ppm}\ \ce{Ni} \tag{6} \end{align}

This means you have $\pu{4.7126 mg}$ nickel per $\pu{1 kg}$ of steel (ppm is one millionth, and thus, mg per kg). Then one tenth of that amount of steel $(\pu{0.1 kg}/\pu{1 kg})$ contains $1/10$ of the $\ce{Ni}$. that is $\pu{0.47126 mg}$

Since it was marked up to $\pu{500 ml}$

In the steel sample: \begin{align} \pu{0.47126 mg}\times\frac{500}{50}&=\pu{4.7126 mg ml-1}\\ \frac{4.7126}{100}&=0.047126\\ \frac{\pu{0.047126 g}}{\pu{0.35 g}}&=0.134645\\ &=13.46\ \%\ \ce{Ni} \tag{7} \end{align}

I feel that I am missing a step in calculations. I would prefer to read a resource of working this out myself than finding the answer. I feel more confident in my second attempt of finding a correct answer.

This question is from a past paper.

$\endgroup$
1
  • 2
    $\begingroup$ Your first attempt goes wrong in your first step; this is corrected in your second attempt, but then you choose to work with 50 mL, which comes out of nowhere. It may help if you first -- ignoring the mathematics -- explain to yourself what you're trying to accomplish in every step and make sure you understand the relevant concepts (proportions and unit conversion). $\endgroup$ Commented Jul 21, 2016 at 13:41

1 Answer 1

2
$\begingroup$

Your first step is incorrect. Think about it this way:

$5$ people have $10$ eyes. This group of people has $14$ eyes. How many people in the group?

The correct solution is:

This group has $14/10 = 1.4$ times more eyes than the known group. Thus, they have $1.4$ times more people. $5$ people $\cdot$($14$ eyes/$10$ eyes) $ = 5$ people $ \times 1.4$ (unitless ratio) $= 7$ people. Here, the units make sense. This is correct.

In your calculation you do ($5$ people)/($10$ eyes $\times 14$ eyes)$= 1$ people/($28$ eyes$^2$). This unit people/eyes$^2$ makes no sense. This is incorrect.

First you find the amount of $\ce{Ni}$ in the final sample.

Your sample absorbs $0.410$ which is slightly less than $0.522$ ($\pu{6 ppm }$sample which). So your sample has $(0.410/0.522) \times \pu{6 ppm} = \pu{4.713 ppm}.$

Water solution has density of $\pu{1 kg/L}$, so $\pu{500 mL} \to \pu{500 g}$. $\pu{4.713 ppm}$ in $\frac{\pu{500 g}}{1\,000\,000} = \pu{0.5 mg}$. $\pu{0.5 mg} \cdot 4.713 = \pu{2.356 mg}$.

The total weight of the sample was $\pu{0.35 g} = \pu{350 mg}$

$$100\ \% \cdot \frac{\pu{2.356 mg}}{\pu{350mg}} = 1.003\ \%$$

You should round it down to $2$ digits, because that was the accuracy of other data. The answer is $1.00\ \%$.

$\endgroup$

Your Answer

By clicking “Post Your Answer”, you agree to our terms of service and acknowledge you have read our privacy policy.

Not the answer you're looking for? Browse other questions tagged or ask your own question.