I know that flash chromatography uses a positive or pushing pressure to achieve the goal which is separation. Would the same goal be achieved if you had a vacuum on the other end instead of a pressure at the top, a vacuum on the bottom?

I have used this technique for ion exchange column with efficient and quick results and wonder if this can be achieved with flash chromatography.

$\endgroup$
  • 1
    $\begingroup$ It should just work the same, but I guess it might limit the solvents you could use (e.g. Et2O might be too volatile) $\endgroup$ – PCK Jun 13 '19 at 20:00
3
$\begingroup$

All chromatography needs to work is some pressure differential, no matter how it's established. Gravitational chromatography uses hydrostatic pressure from the weight of the solvent, and a typical gravitational column can have a roughly ~20-50 mbar pressure difference between the top and bottom of the column; flash chromatography also has the hydrostatic pressure, but has an additional positive gas pressure at the top from e.g. pumped air, and can typically reach ~100 mbar pressure differences.

The principle of chromatography is to force compounds in a mixture to stay at the interface of a stationary phase (e.g. silica) and a mobile phase. The slight differences in how molecules behave in this boundary are added up over time, creating the observed separation.

Since this is a surface process, having a stationary phase with higher specific surface area and better packing effectively means molecules spend more time undergoing separation at the interface than being dissolved in the bulk of the eluent, where no separation happens. However, a more tightly packed stationary phase also restricts eluent flow. Under a simplified set of assumptions, decreasing the stationary phase particle size by half (and therefore increasing specific surface area by a factor of 4) requires 4 times the pressure differential to produce the same flow rate. So getting better separation in reasonable timeframes requires larger pressure differentials.

This is where vacuum chromatography techniques come in. Vacuum pumps are a staple laboratory instrument, and can easily achieve negative pressures of ~900 mbar with common glassware. Therefore, without too much hassle, it is possible to obtain good flow rates using finer stationary phase particle sizes and get improved separation.

Of course, larger pressure differentials can be achieved with positive pressure at the top of the column. Medium, high and ultra-high pressure chromatography (MPLC, HPLC, UHPLC) make use of stationary phases with very fine particles (from ~10 μm to ~1 μm diameter) which start to look more like silt or clay. The only way to force liquids through these stationary phases is with very large pressure differentials, from ~5 bar to as much as ~1000 bar, but they can achieve incredible resolution of compounds during separation. The reason UHPLC isn't used for every single separation is cost and scalability; it's an engineering challenge to control such enormous pressures safely. Glass is very resilient under compression (which is why glassware can commonly hold a vacuum), but weak under expansion. This means such positive pressures need much more rigid containers, such as thick steel columns. This also limits their feasible sizes and therefore the amount of stationary phase/eluent/compound they can handle.

As an example of a vacuum chromatography technique, I point towards the excellent video by Daniel Pedersen detailing dry column vacuum chromatography (DCVC). There is also a preceding publication which goes into further details. For what it's worth, I've found this vacuum chromatography technique to be far superior to flash chromatography.

| improve this answer | |
$\endgroup$
2
$\begingroup$

In order to make the fluid flow, you just need the pressure differential. At a given flow rate, the head pressure (at the top) and bottom pressure can be any number as long as the difference in pressure is the same.

$$ \begin{align} P_\mathrm{head} &= \pu{1.0 atm} \\ P_\mathrm{end} &= \pu{0.1 atm} \end{align} $$

will have the same flow rate as

$$ \begin{align} P_\mathrm{head} &= \pu{1.9 atm} \\ P_\mathrm{end} &= \pu{1.0 atm} \end{align} $$

Mechanically it is more convenient to generate pressure rather than generate vacuum.

| improve this answer | |
$\endgroup$
  • 1
    $\begingroup$ At low absolute pressure, you have a much larger danger of rupture of the static phase of your column. Especially during startup, the form of the pressure gradient in the column is likely fatal. $\endgroup$ – Karl Jun 13 '19 at 21:01
0
$\begingroup$

Very detailed answer by Nicolau Neto. I wish to add a clarifiction "The reason UHPLC isn't used for every single separation is cost and scalability; it's an engineering challenge to control such enormous pressures safely. " is not entirely correct. I have had good discussions with the person who invented the ultrahigh pressure liquid chromatograph (UHPLC). Although 1000 bar may sound dangerous, the actual force to generate this high pressure very small.* In a UHPLC the typical flow rate is on the order of 0.2 to 2 mL/min (max). Higher end or the latest ones have upto 8 mL/min. All they do is make the piston diameter smaller. Recall that pressure =force/unit area. If the cross-sectional area of the pump head bore and the piston bore decreases but the force pushing the pistons remains the same, then you have a UHPLC pump. The conceptual issues with a UHPLC is not the safety concern at all, but frictional heating with very small particles, heating of the mobile phase during compression/expansion, and leakages at pressure extremes, at every possible tubing connection. Eventually, many companies started to use gold seals, as well as gold coated connections. Secondly, because of viscous frictional heating, narrow columns such as 1 to 2.1 mm i.d. must be used on a UHPLC. Radial temperature gradients inside a packed bed ruins the separation.


  • A large pump such as a pneumatic pump at 1000 bar is very very dangerous. You need a lot of safety measures. These pumps are used for packing very small particles in stainless steel columns.